- pretty_name: "OpenH-RF fullwave-abdominal-wall: Fullwave Abdominal Wall Simulation (UNC / NC State / Stanford)"
 
3
  license: cc-by-4.0
4
  task_categories:
5
  - image-segmentation
@@ -18,32 +19,26 @@ size_categories:
18
  - 1K<n<10K
19
  ---
20
 
21
- # fullwave-abdominal-wall: Fullwave abdominal wall simulation dataset
22
 
23
- ## Dataset Description
 
 
24
 
25
- Full-wave nonlinear acoustic simulations of transabdominal liver imaging through
26
- anatomically realistic numerical phantoms of the human abdominal wall. Each acquisition
27
- pairs **full synthetic aperture (multistatic) RF channel data** with the **exact
28
- ground-truth material maps that generated it**: speed of sound, density, absorption and
29
- the coefficient of nonlinearity, plus a per-pixel tissue segmentation.
30
 
31
- The phantoms were segmented from Visible Human Project cryosection photography at
32
- 0.33 mm isotropic resolution and interpolated to 0.0825 mm. Wave propagation was
33
- simulated with Fullwave 2, which models reverberation, aberration and nonlinear
34
- propagation. The transducer modelled is a curvilinear C5-2v (Verasonics).
35
 
36
- This is **simulated** data. Its distinguishing property is that the ground truth is
37
- exact rather than estimated: the sound-speed and attenuation maps are the simulation
38
- inputs, not a reconstruction, which makes it directly usable for training and
39
- quantitative evaluation of sound-speed estimation and aberration-correction methods.
40
 
41
- ## Dataset Contributors
42
 
43
- University of North Carolina at Chapel Hill, NC State University, and Stanford University.
44
 
45
- Primary point of contact: **Gianmarco Pinton** (gia@email.unc.edu), Lampe Joint
46
- Department of Biomedical Engineering, UNC Chapel Hill and NC State University.
 
 
47
 
48
  The underlying phantoms and simulations are described in Zhuang et al. (2026).
49
 
@@ -53,81 +48,55 @@ The underlying phantoms and simulations are described in Zhuang et al. (2026).
53
 
54
  ## License / Terms of Use
55
 
56
- **CC BY 4.0** (see `LICENSE`).
 
 
 
 
 
 
 
 
 
 
 
 
 
 
 
 
57
 
58
- The material distributed here (the RF channel data, the acoustic material maps, and the
59
- segmentations) is original scholarly output of the contributing institutions
60
- (simulations and phantom construction by the UNC / Stanford team) and is released under
61
- CC BY 4.0.
62
 
63
- The phantoms derive from the **Visible Human Project** (VHP) cryosection imagery of the
64
- U.S. National Library of Medicine. As of July 2019 the NLM Data License was replaced by
65
- open Terms and Conditions
66
- (<https://www.nlm.nih.gov/databases/download/terms_and_conditions.html>): no license
67
- agreement or registration is required, there is no restriction on commercial use, no
68
- royalties, and no share-alike or redistribution restriction, so nothing in the VHP
69
- terms conflicts with CC BY 4.0. NLM works are U.S. Government works and carry no
70
- copyright in the United States.
71
 
72
  The one obligation carried over from the VHP terms is acknowledgement:
73
 
74
  > **Courtesy of the U.S. National Library of Medicine.**
75
 
76
- The NLM does not endorse this dataset, and nothing here should be read as implying such
77
- endorsement.
78
 
79
- *This is a good-faith reading of the public terms, not legal advice; the contributor is
80
- the responsible party for the license declaration.*
81
 
82
- ## Intended Usage
83
 
84
- - **Speed-of-sound estimation:** exact per-pixel ground-truth sound speed, randomised
85
- per tissue per simulation so the mapping is not memorisable from anatomy alone.
86
- - **Aberration correction:** measured RMS arrival-time aberration is
87
- 138.9 ± 78.4 ns across the dataset.
88
- - **Reverberation / clutter suppression:** the abdominal wall produces realistic
89
- diffuse reverberation (decay in the expected −12 to −10 dB/cm range).
90
- - **Advanced beamforming:** the multistatic full synthetic aperture matrix supports
91
- retrospective synthesis of any transmit sequence (focused, plane wave, diverging).
92
- - **Tissue segmentation** from channel data.
93
 
94
- ## Dataset Characterization
95
-
96
- - **Data Collection Method:** synthetic (Fullwave 2 numerical simulation)
97
- - **Labeling Method:** synthetic ground truth (simulation inputs)
98
- - **Acquisition system:** simulated Verasonics **C5-2v curvilinear array**: 128
99
- elements, 49.57 mm radius of curvature, 0.508 mm arc pitch, 3.7 MHz transmit centre
100
- frequency, 70% fractional bandwidth, 14.436 MHz sampling. Transmit pulse is a
101
- two-cycle Gaussian-enveloped sine at 0.1 MPa. No lens or matching layer is modelled.
102
 
103
  ## Dataset Format
104
 
105
- All files are in the *zea* file format (current release `zea_version` 0.1.4), one HDF5 file per
106
- acquisition, single track.
107
-
108
- **Pre-processing applied before packaging:** none to the channel data. It is the
109
- native simulator output at full rate. The simulation ran at 101.1 MHz and was
110
- decimated by 7 to the stored 14.436 MHz *by the simulator*, before this packaging.
111
-
112
- Coordinate frame: `x` is lateral, `y ≡ 0` (the simulations are 2-D), `z` is depth.
113
- The **array apex is at z = 0** and the centre of curvature at `z = −49.57 mm`, matching
114
- zea's `polar_pixel_grid` convention. `probe_geometry`, `transmit_origins` and every map
115
- `coordinates` array share this frame, so the channel data and the maps are exactly
116
- registered.
117
-
118
- Transmit sequence: **full synthetic aperture**: each element fires alone and all 128
119
- elements receive. Hence `t0_delays` is all-zero, `tx_apodizations` is the identity, and
120
- `focus_distances` is zero. `initial_times` is **negative** (−79.17 ns): the simulator's
121
- transmit start time offset means sample 0 corresponds to a two-way time of flight of
122
- −t0.
123
-
124
- Transmit and receive element positions differ slightly. `probe_geometry` holds the
125
- **receive** element positions and `scan/transmit_origins` the **transmit** element
126
- positions; the Fullwave simulation discretizes the transmit and receive apertures onto
127
- separate arcs of its grid that sit about 0.16 mm (≈ 0.4 wavelengths) apart. This is a
128
- property of the source simulation, not an error. When beamforming, use
129
- `transmit_origins` for the transmit leg and `probe_geometry` for the receive leg, as the
130
- shipped reference reconstruction does.
131
 
132
  ### Per-sample feature table
133
 
@@ -155,8 +124,7 @@ shipped reference reconstruction does.
155
  | `scan/polar_angles` | (128,) | float32 | rad | Element normal angle, ±0.6507 |
156
  | `probe/probe_geometry` | (128, 3) | float32 | m | Receive element positions on the arc |
157
 
158
- All maps are on the polar reconstruction grid: 640 radial samples spanning 5.0–71.5 mm
159
- from the array surface, by 128 beams spanning ±0.3 rad.
160
 
161
  ## Dataset Quantification
162
 
@@ -164,21 +132,11 @@ from the array surface, by 128 beams spanning ±0.3 rad.
164
 
165
  - **1906 acquisitions**, 1 frame each, 128 transmits × 128 receives per acquisition.
166
  - **Stored HDF5 size:** 202.69 GB (202,692,755,456 bytes).
167
- - Drawn from **14 distinct phantom volumes** derived from a single Visible Human
168
- subject: `vishuman_abdominal_cropped` (308), `vishuman_abdwall_153_h` (187),
169
- `vishuman_abdwall_cropped_7-153s` (148), `vishuman_abdwall_cropped_5left153s` (134),
170
- `vishuman_abdwall_cropped_7left153s` (133), `vishuman_abdwall_set_04` (123),
171
- `vishuman_abdwall_cropped_6left153s` (119), `vishuman_abdwall_set_13` (114),
172
- `vishuman_abdwall_set_05` (113), `vishuman_abdwall_set_12` (109),
173
- `vishuman_abdwall_set_07` (109), `vishuman_abdwall_set_10` (108),
174
- `vishuman_abdwall_set_06` (104), `vishuman_abdwall_set_11` (97).
175
 
176
  ### Splits
177
 
178
- `dataset_split.csv` ships alongside the HDF5 files. Splits are **grouped by phantom
179
- volume** so that no volume appears in both train and validation. This avoids anatomical
180
- leakage, which matters because many acquisitions are different 2-D slices of the same
181
- volume.
182
 
183
  | Split | Train | Validation |
184
  |---|---|---|
@@ -189,88 +147,37 @@ volume.
189
  | `fold3` | 1314 | 488 |
190
  | `fold4` | 1328 | 465 |
191
 
192
- Note the `fold*` splits are **not** complementary: between 97 and 153 acquisitions are
193
- in neither the train nor the validation set of a given fold.
194
 
195
  ## Subject Metadata
196
 
197
- Aggregate only; no PHI. All phantoms derive from **one** Visible Human Project cadaveric
198
- subject (female), so the dataset represents a single anatomy sampled at many slice
199
- positions and orientations, with randomised abdominal wall thickness (5–71 mm, mean
200
- 41.0 mm, SD 8.9 mm) and randomised per-tissue acoustic properties.
201
 
202
- Tissue composition varies by volume type; across the phantom volumes reported in the
203
- source publication, subcutaneous fat and muscle predominate (fat 56.5 ± 8.3%, muscle
204
- 22.2 ± 7.7%, connective 20.7 ± 4.2%, blood 0.6 ± 0.5% for wall-dominated volumes).
205
 
206
  ## Data Validation
207
 
208
- `reconstruct.py` builds a `zea.Pipeline`
209
- (`Cast → Demodulate → Beamform(delay_and_sum) → EnvelopeDetect → Normalize → LogCompress`),
210
- reconstructs from the raw channel data on the polar grid, scan-converts to a physical
211
- sector and writes a PNG. The pipeline plus its grid parameters are saved in
212
- `pipeline.yaml`.
213
-
214
- For current release files, use `zea==0.1.6` and any Keras backend (`reconstruct.py` falls back to torch if
215
- `KERAS_BACKEND` is unset, but respects whatever you have configured).
216
-
217
- ```bash
218
- python reconstruct.py <file>.hdf5 --out bmode.png --save-yaml pipeline.yaml
219
- ```
220
 
221
  Reference outputs: `bmode_reference.png`.
222
 
223
- **Geometry validation.** The zea reconstruction was checked against the reference
224
- delay-and-sum image shipped with the source dataset: radial registration lag **0
225
- samples**, lateral lag **0 beams**, structural correlation 0.92–0.95 (8×8 and 16×16
226
- smoothed). Residual speckle-level decorrelation is expected: the reference used a
227
- Hilbert analytic signal with bicubic interpolation, while the zea pipeline demodulates
228
- to baseband with its own interpolation and apodization.
229
 
230
  ## Known Issues
231
 
232
- - **`fs / f0` = 14.436 / 3.7 = 3.90**, marginally below the ×4 heuristic often applied
233
- to RF. This is the simulator's native rate (101.1 MHz decimated by 7) and still
234
- satisfies Nyquist for the −6 dB bandwidth of a 70%-fractional-bandwidth 3.7 MHz pulse
235
- (band edge ≈ 5.0 MHz, Nyquist 7.2 MHz). No aliasing is present.
236
- - **`segmentation` is derived, not stored.** The simulator's own integer segmentation is
237
- a *per-volume region index* whose index-to-tissue mapping differs between phantom
238
- volumes, so those indices are not comparable across acquisitions. It is preserved
239
- verbatim as `phantom_region_map` for provenance. The shipped `segmentation` is
240
- recovered per pixel from the exact (density, β, absorption) fingerprint of each tissue
241
- in the simulator's material table. These three properties are constants, whereas
242
- sound speed is randomly drawn per tissue per simulation and is therefore excluded from
243
- the classification. Pixels on tissue boundaries are softened by the simulator's
244
- Gaussian blur (σ = 1 pixel) and are assigned to the nearest pure tissue.
245
- - **2-D simulations.** Elevational focusing and out-of-plane scattering are not
246
- modelled. The source publication argues this is acceptable for subcostal
247
- transabdominal scanning, where ribs do not obstruct the field.
248
- - **Single subject.** All anatomy derives from one Visible Human cadaver, so anatomical
249
- variability is limited to slice position, orientation, wall-thickness scaling and
250
- randomised acoustic properties. Simulations overlap spatially due to random sampling
251
- within volumes; the `volume` column allows filtering for overlap.
252
- - **Postmortem blood redistribution** in the supine source cadaver produces additional
253
- contrast of connective tissue in the posterior portions of the images and limits
254
- connective tissue visibility anteriorly.
255
 
256
  ## Ethical Considerations
257
 
258
- The data is entirely synthetic. It contains no patient data, no PHI and no identifiable
259
- information. The underlying anatomy derives from the **Visible Human Project**, a
260
- publicly released cadaveric imaging dataset collected with documented donor consent by
261
- the U.S. National Library of Medicine; no living subjects are involved and no IRB
262
- approval is applicable to the simulation work. The VHP source imagery is provided under
263
- open NLM Terms and Conditions and is acknowledged as required (see License / Terms of
264
- Use). No patient consent or de-identification concerns arise because no patient data is
265
- present at any stage.
266
 
267
  ## Citation
268
 
269
- > L. Zhuang, O. Ostras, M. Sode, W. Simson, D. Hyun, F. Santibanez, J. Dahl and
270
- > G. Pinton, "Labeled Numerical Phantom of Abdominal Wall for Wave-Physics-Based
271
- > Ultrasound Imaging: Applications to Image Reconstruction," *IEEE Transactions on
272
- > Ultrasonics, Ferroelectrics, and Frequency Control*, vol. 73, no. 1, pp. 24–34,
273
- > Jan. 2026. doi:10.1109/TUSON.2025.3638314
274
 
275
  Related resources:
276
 
 
1
  ---
2
+ name: unc-liver
3
+ pretty_name: "fullwave-abdominal-wall: Fullwave Abdominal Wall Simulation (UNC / NC State / Stanford)"
4
  license: cc-by-4.0
5
  task_categories:
6
  - image-segmentation
 
19
  - 1K<n<10K
20
  ---
21
 
22
+ # fullwave-abdominal-wall: Fullwave Abdominal Wall Simulation Dataset
23
 
24
+ ![Scan-converted B-mode through a simulated abdominal wall](assets/main.png)
25
+
26
+ *Scan-converted B-mode reconstructed from the full-synthetic-aperture channel data of one simulated acquisition in [`data/`](https://huggingface.co/datasets/nvidia/OpenH-RF/tree/main/unc-liver/data): the abdominal wall layers in the near field above the liver.*
27
 
28
+ ## Dataset Description
 
 
 
 
29
 
30
+ Full-wave nonlinear acoustic simulations of transabdominal liver imaging through anatomically realistic numerical phantoms of the human abdominal wall. Each acquisition pairs **full synthetic aperture (multistatic) RF channel data** with the **exact ground-truth material maps that generated it**: speed of sound, density, absorption and the coefficient of nonlinearity, plus a per-pixel tissue segmentation.
 
 
 
31
 
32
+ The phantoms were segmented from Visible Human Project cryosection photography at 0.33 mm isotropic resolution and interpolated to 0.0825 mm. Wave propagation was simulated with Fullwave 2, which models reverberation, aberration and nonlinear propagation. The transducer modelled is a curvilinear C5-2v (Verasonics).
 
 
 
33
 
34
+ This is **simulated** data. Its distinguishing property is that the ground truth is exact rather than estimated: the sound-speed and attenuation maps are the simulation inputs, not a reconstruction, which makes it directly usable for training and quantitative evaluation of sound-speed estimation and aberration-correction methods.
35
 
36
+ ## Dataset Contributor(s)
37
 
38
+ - Gianmarco Pinton <gia@email.unc.edu> (primary point of contact; Lampe Joint Department of Biomedical Engineering, UNC Chapel Hill and NC State University)
39
+ - University of North Carolina at Chapel Hill
40
+ - NC State University
41
+ - Stanford University
42
 
43
  The underlying phantoms and simulations are described in Zhuang et al. (2026).
44
 
 
48
 
49
  ## License / Terms of Use
50
 
51
+ [Creative Commons Attribution 4.0 International (CC BY 4.0)](https://creativecommons.org/licenses/by/4.0/legalcode.en). Retain attribution and identify modifications when reusing the data.
52
+
53
+ ## Intended Usage
54
+
55
+ - **Speed-of-sound estimation:** exact per-pixel ground-truth sound speed, randomised per tissue per simulation so the mapping is not memorisable from anatomy alone.
56
+ - **Aberration correction:** measured RMS arrival-time aberration is 138.9 ± 78.4 ns across the dataset.
57
+ - **Reverberation / clutter suppression:** the abdominal wall produces realistic diffuse reverberation (decay in the expected −12 to −10 dB/cm range).
58
+ - **Advanced beamforming:** the multistatic full synthetic aperture matrix supports retrospective synthesis of any transmit sequence (focused, plane wave, diverging).
59
+ - **Tissue segmentation** from channel data.
60
+
61
+ ## Dataset Characterization
62
+
63
+ - **Data Collection Method:** synthetic (Fullwave 2 numerical simulation)
64
+ - **Labeling Method:** synthetic ground truth (simulation inputs)
65
+ - **Acquisition system:** simulated Verasonics **C5-2v curvilinear array**: 128 elements, 49.57 mm radius of curvature, 0.508 mm arc pitch, 3.7 MHz transmit centre frequency, 70% fractional bandwidth, 14.436 MHz sampling. Transmit pulse is a two-cycle Gaussian-enveloped sine at 0.1 MPa. No lens or matching layer is modelled.
66
+
67
+ ## Source Attribution
68
 
69
+ The material distributed here (the RF channel data, the acoustic material maps, and the segmentations) is original scholarly output of the contributing institutions (simulations and phantom construction by the UNC / Stanford team) and is released under CC BY 4.0.
 
 
 
70
 
71
+ The phantoms derive from the **Visible Human Project** (VHP) cryosection imagery of the U.S. National Library of Medicine. As of July 2019 the NLM Data License was replaced by open Terms and Conditions (<https://www.nlm.nih.gov/databases/download/terms_and_conditions.html>): no license agreement or registration is required, there is no restriction on commercial use, no royalties, and no share-alike or redistribution restriction, so nothing in the VHP terms conflicts with CC BY 4.0. NLM works are U.S. Government works and carry no copyright in the United States.
 
 
 
 
 
 
 
72
 
73
  The one obligation carried over from the VHP terms is acknowledgement:
74
 
75
  > **Courtesy of the U.S. National Library of Medicine.**
76
 
77
+ The NLM does not endorse this dataset, and nothing here should be read as implying such endorsement.
 
78
 
79
+ *This is a good-faith reading of the public terms, not legal advice; the contributor is the responsible party for the license declaration.*
 
80
 
81
+ ## Processing the Dataset
82
 
83
+ The acquisitions can be processed with the `reconstruct.py` [script](https://github.com/open-h/OpenH-RF/blob/main/datasets/unc-liver/reconstruct.py) as provided in the [OpenH-RF GitHub repository](https://github.com/open-h/OpenH-RF), together with the `pipeline.yaml` definition in this folder and the [zea library](https://github.com/tue-bmd/zea). The script streams the data from the Hugging Face Hub.
 
 
 
 
 
 
 
 
84
 
85
+ Set `ZEA_FILE` and `FRAME` at the top of the script to pick an acquisition. For current release files, use `zea==0.1.6` and any Keras backend (`reconstruct.py` falls back to torch if `KERAS_BACKEND` is unset, but respects whatever you have configured).
 
 
 
 
 
 
 
86
 
87
  ## Dataset Format
88
 
89
+ [zea v0.1.4](https://github.com/tue-bmd/zea)
90
+
91
+ All files are in the *zea* file format (current release `zea_version` 0.1.4), one HDF5 file per acquisition, single track.
92
+
93
+ **Pre-processing applied before packaging:** none to the channel data. It is the native simulator output at full rate. The simulation ran at 101.1 MHz and was decimated by 7 to the stored 14.436 MHz *by the simulator*, before this packaging.
94
+
95
+ Coordinate frame: `x` is lateral, `y ≡ 0` (the simulations are 2-D), `z` is depth. The **array apex is at z = 0** and the centre of curvature at `z = −49.57 mm`, matching zea's `polar_pixel_grid` convention. `probe_geometry`, `transmit_origins` and every map `coordinates` array share this frame, so the channel data and the maps are exactly registered.
96
+
97
+ Transmit sequence: **full synthetic aperture**: each element fires alone and all 128 elements receive. Hence `t0_delays` is all-zero, `tx_apodizations` is the identity, and `focus_distances` is zero. `initial_times` is **negative** (−79.17 ns): the simulator's transmit start time offset means sample 0 corresponds to a two-way time of flight of −t0.
98
+
99
+ Transmit and receive element positions differ slightly. `probe_geometry` holds the **receive** element positions and `scan/transmit_origins` the **transmit** element positions; the Fullwave simulation discretizes the transmit and receive apertures onto separate arcs of its grid that sit about 0.16 mm (≈ 0.4 wavelengths) apart. This is a property of the source simulation, not an error. When beamforming, use `transmit_origins` for the transmit leg and `probe_geometry` for the receive leg, as the shipped reference reconstruction does.
 
 
 
 
 
 
 
 
 
 
 
 
 
 
 
100
 
101
  ### Per-sample feature table
102
 
 
124
  | `scan/polar_angles` | (128,) | float32 | rad | Element normal angle, ±0.6507 |
125
  | `probe/probe_geometry` | (128, 3) | float32 | m | Receive element positions on the arc |
126
 
127
+ All maps are on the polar reconstruction grid: 640 radial samples spanning 5.0–71.5 mm from the array surface, by 128 beams spanning ±0.3 rad.
 
128
 
129
  ## Dataset Quantification
130
 
 
132
 
133
  - **1906 acquisitions**, 1 frame each, 128 transmits × 128 receives per acquisition.
134
  - **Stored HDF5 size:** 202.69 GB (202,692,755,456 bytes).
135
+ - Drawn from **14 distinct phantom volumes** derived from a single Visible Human subject: `vishuman_abdominal_cropped` (308), `vishuman_abdwall_153_h` (187), `vishuman_abdwall_cropped_7-153s` (148), `vishuman_abdwall_cropped_5left153s` (134), `vishuman_abdwall_cropped_7left153s` (133), `vishuman_abdwall_set_04` (123), `vishuman_abdwall_cropped_6left153s` (119), `vishuman_abdwall_set_13` (114), `vishuman_abdwall_set_05` (113), `vishuman_abdwall_set_12` (109), `vishuman_abdwall_set_07` (109), `vishuman_abdwall_set_10` (108), `vishuman_abdwall_set_06` (104), `vishuman_abdwall_set_11` (97).
 
 
 
 
 
 
 
136
 
137
  ### Splits
138
 
139
+ `dataset_split.csv` ships alongside the HDF5 files. Splits are **grouped by phantom volume** so that no volume appears in both train and validation. This avoids anatomical leakage, which matters because many acquisitions are different 2-D slices of the same volume.
 
 
 
140
 
141
  | Split | Train | Validation |
142
  |---|---|---|
 
147
  | `fold3` | 1314 | 488 |
148
  | `fold4` | 1328 | 465 |
149
 
150
+ Note the `fold*` splits are **not** complementary: between 97 and 153 acquisitions are in neither the train nor the validation set of a given fold.
 
151
 
152
  ## Subject Metadata
153
 
154
+ Aggregate only; no PHI. All phantoms derive from **one** Visible Human Project cadaveric subject (female), so the dataset represents a single anatomy sampled at many slice positions and orientations, with randomised abdominal wall thickness (5–71 mm, mean 41.0 mm, SD 8.9 mm) and randomised per-tissue acoustic properties.
 
 
 
155
 
156
+ Tissue composition varies by volume type; across the phantom volumes reported in the source publication, subcutaneous fat and muscle predominate (fat 56.5 ± 8.3%, muscle 22.2 ± 7.7%, connective 20.7 ± 4.2%, blood 0.6 ± 0.5% for wall-dominated volumes).
 
 
157
 
158
  ## Data Validation
159
 
160
+ `reconstruct.py` builds a `zea.Pipeline` (`Cast → Demodulate → Beamform(delay_and_sum) → EnvelopeDetect → Normalize → LogCompress`), reconstructs from the raw channel data on the polar grid, scan-converts to a physical sector and writes a PNG. The pipeline plus its grid parameters are saved in `pipeline.yaml`.
 
 
 
 
 
 
 
 
 
 
 
161
 
162
  Reference outputs: `bmode_reference.png`.
163
 
164
+ **Geometry validation.** The zea reconstruction was checked against the reference delay-and-sum image shipped with the source dataset: radial registration lag **0 samples**, lateral lag **0 beams**, structural correlation 0.92–0.95 (8×8 and 16×16 smoothed). Residual speckle-level decorrelation is expected: the reference used a Hilbert analytic signal with bicubic interpolation, while the zea pipeline demodulates to baseband with its own interpolation and apodization.
 
 
 
 
 
165
 
166
  ## Known Issues
167
 
168
+ - **`fs / f0` = 14.436 / 3.7 = 3.90**, marginally below the ×4 heuristic often applied to RF. This is the simulator's native rate (101.1 MHz decimated by 7) and still satisfies Nyquist for the −6 dB bandwidth of a 70%-fractional-bandwidth 3.7 MHz pulse (band edge ≈ 5.0 MHz, Nyquist 7.2 MHz). No aliasing is present.
169
+ - **`segmentation` is derived, not stored.** The simulator's own integer segmentation is a *per-volume region index* whose index-to-tissue mapping differs between phantom volumes, so those indices are not comparable across acquisitions. It is preserved verbatim as `phantom_region_map` for provenance. The shipped `segmentation` is recovered per pixel from the exact (density, β, absorption) fingerprint of each tissue in the simulator's material table. These three properties are constants, whereas sound speed is randomly drawn per tissue per simulation and is therefore excluded from the classification. Pixels on tissue boundaries are softened by the simulator's Gaussian blur (σ = 1 pixel) and are assigned to the nearest pure tissue.
170
+ - **2-D simulations.** Elevational focusing and out-of-plane scattering are not modelled. The source publication argues this is acceptable for subcostal transabdominal scanning, where ribs do not obstruct the field.
171
+ - **Single subject.** All anatomy derives from one Visible Human cadaver, so anatomical variability is limited to slice position, orientation, wall-thickness scaling and randomised acoustic properties. Simulations overlap spatially due to random sampling within volumes; the `volume` column allows filtering for overlap.
172
+ - **Postmortem blood redistribution** in the supine source cadaver produces additional contrast of connective tissue in the posterior portions of the images and limits connective tissue visibility anteriorly.
 
 
 
 
 
 
 
 
 
 
 
 
 
 
 
 
 
 
173
 
174
  ## Ethical Considerations
175
 
176
+ The data is entirely synthetic. It contains no patient data, no PHI and no identifiable information. The underlying anatomy derives from the **Visible Human Project**, a publicly released cadaveric imaging dataset collected with documented donor consent by the U.S. National Library of Medicine; no living subjects are involved and no IRB approval is applicable to the simulation work. The VHP source imagery is provided under open NLM Terms and Conditions and is acknowledged as required (see License / Terms of Use). No patient consent or de-identification concerns arise because no patient data is present at any stage.
 
 
 
 
 
 
 
177
 
178
  ## Citation
179
 
180
+ > L. Zhuang, O. Ostras, M. Sode, W. Simson, D. Hyun, F. Santibanez, J. Dahl and G. Pinton, "Labeled Numerical Phantom of Abdominal Wall for Wave-Physics-Based Ultrasound Imaging: Applications to Image Reconstruction," *IEEE Transactions on Ultrasonics, Ferroelectrics, and Frequency Control*, vol. 73, no. 1, pp. 24–34, Jan. 2026. doi:10.1109/TUSON.2025.3638314
 
 
 
 
181
 
182
  Related resources: